Gradient Verification for Biological Separations Using the VERITY® 1810 Conductivity and pH Monitor
페이스북 트위터 구글 싸이월드 마이피플 
작성자 : 사이언스21 | 조회: 41 | 날짜: 2020-09-03 10:01:42


INTRODUCTION
Conductivity and /or pH monitoring are commonly used for biological separations and is necessary for the purification of oligonucleotides, monoclonal antibodies(mAbs), and other proteins. The VERITY® 1810 Conductivity and pH Monitor is a state-of-the-art detection instrument for biological purifications with a large dynamic range and rapid response time for real-time gradient monitoring.

This technical note highlights the ability of the VERITY® 1810 Conductivity and pH Monitor under TRILUTION® LC control to verify gradient composition during a chromatographic run.

Figure1. VERITY® 1810 Conductivity and pH Monitor with Optional pH Flow Cell.


Materials & Methods
Standard Solutions
For conductivity monitoring, a solution of 100 mM phosphate buffer was prepared by dissolving powdered reagents in MilliQ purified water. MilliQ purified water was used as a solvent for the conductivity gradient.
For pH monitoring, solutions of 20 mM phosphate buffer and 0.1 M citric acid were prepared by dissolving powdered reagents in MilliQ purified water.

Instrumentation and Software
VERITY® 1810 Conductivity and pH Monitor with optional pH flow cell
•322 HPLC Pump (H1 heads)
•Minipuls®3 (for washing the pump heads)
•TRILUTION® LC

Conductivity Monitoring Methods
This experiment simulated use of a size exclusion or ion exchange column running a step gradient(Table 1) to mimic protein separation via buffers of varying ionic strengths.
This experiment was conducted at room uploaderature and in a refrigerated unit (~10°C).

Table 1. Conductivity Monitoring Gradient

pH Monitoring Methods
The experiment simulated the use of a 1mL protein column to run a buffer exchange, loading with five column volumes of binding buffer, eluting with five column volumes of elution buffer, and equilibrating with ten column volumes of binding buffer(Table 2).
The VERITY® 1810 Conductivity and pH Monitor was used to monitor the mobile phase pH change with buffer composition change.
This experiment was run at room uploaderature and in a refrigerated unit (~12°C).

Table 2. pH Monitoring Gradient

Figure 2. Example TRILUTION® LC method for conductivity monitoring

Figure 3. Example TRILUTION® LC method for pH monitoring


Results and Discussion
Conductivity Monitoring
At both room uploaderature and at 10°C, the VERITY® 1810 Conductivity and pH Monitor effectively monitored the changes to the mobile phase composition between 0% and 100% phosphate buffer(Figure 4).
As the percentage of phosphate buffer increased, the VERITY® 1810 Monitor exhibited a corresponding increase in the conductivity signal. The VERITY® 1810 Monitor also indicated the return of the flow to effectively 0mS/cm with the transition to 0% phosphate buffer.
The results were similar at room uploaderature and at 10°C, demonstrating the ability of the VERITY® 1810 Monitor to collect conductivity data in different uploaderature environments without requiring recalibration.
Conductivity and uploaderature data were collected simultaneously during the run(data not shown).

Figure 4. Conductivity monitoring at room uploaderature(Aqua, Pink, Purple) and at 10°C(Blue, Orange, Green)in relation to mobile phase concentration(Dark red).

pH Monitoring
At both room uploaderature and at 12°C, the VERITY® 1810 Conductivity and pH Monitor effectively monitored the changes to the mobile phase composition, transitioning between 100% 20 mM phosphate buffer (pH 8) and 100% 0.1M citric acid (pH 2), and then returning to 100% 20 mM
phosphate buffer (pH 8) (Figure 5). The VERITY® 1810 Monitor accurately followed the changes in pH, showing that the monitor can be used to detect the pH of the system.
Results at room uploaderature were similar to results at lowered uploaderature, showing the ability of the VERITY® 1810 Monitor to collect pH data in different uploaderature environments. pH, conductivity, and data were collected simultaneously during the run (data not shown).

Figure 5. pH results at room uploaderature(Aqua, Red, Purple) and at 12°C(Blue, Orange, Green)in relation to mobile phase concentration(Dark red)

Summary
With the Gilson VERITY® 1810 Conductivity and pH Monitor, real-time conductivity, pH, and uploaderature signals are monitored efficiently andNeffectively, which allows the user to protect their valuable samples, ensure product purity, and optimize the purification process.

Rerference(참고문헌): Gilson Technical note TRANS0616
Model Name(모델명): VERITY®1810 Conductivity and pH Monitor
The Person in Charge(담당자): An Hyesook
Maker(제조사): Gilson Inc.
Country of Origin(원산지): U.S.A
e-mail: kaisco1@kaisco.co.kr
Data Services(자료제공): Gilson Inc.


댓글 0
 

댓글쓰기
스팸방지를 위하여 보이는 순서대로 모두 입력해 주세요.
이미지 새로고침
이전글 Application Note•q TOWER³ 84 사이언스21 2020-09-03 09:51

비밀번호 확인

댓글 등록시에 입력했던 비밀번호를 입력해 주세요.

비밀번호 확인

게시글 등록시에 입력했던 비밀번호를 입력해 주세요.

번호제목작성자등록일조회수
>> Gradient Verification for Biological Separations U... 사이언스212020.09.0342
28 Application Note•q TOWER³ 84 사이언스212020.09.0314
27 GX-271 Oligo Purification System: Streamlined Puri... 사이언스212020.07.0897
26 Application Note•q TOWER³ 사이언스212020.07.0899
25 참치 중 수은의 연속 측정 사이언스212020.07.0882
24 Automation of RT-qPCR for Promega’s GoTaq® Probe 1... 사이언스212020.06.10110
23 SARS-CoV-2 사이언스212020.04.16108
22 고품질, 합리적 가격, 편리한 Operation “ Chemstudio series ” 사이언스212020.03.05183
21 96,384 and 1536 Channel simultaneous pipettor “ Cy... 사이언스212020.02.07230
20 Gene Expression Plant - Virus Study Monitoring Pot... 사이언스212019.11.06100
19 High Quality, High Precision real time PCR “ q TOW... 사이언스212019.11.06137
18 참치 중 수은의 연속 분석을 통해 알아본 수은 분석기 수산물 응용 사례 사이언스212019.10.10179
17 멀티채널 피펫팅 시스템을 이용한 Coomassie(Bradford) Protein Assa... 사이언스212019.09.03406
16 Monitoring Drinking Water Quality with the PlasmaQ... 사이언스212018.12.05301
15 제약 샘플 분석을 위한 Rigaku장비 소개 사이언스212018.07.12589
14 생리활성화합물의 작용기전 연구를 위한 융합기술(Integrative Technology f... 사이언스212018.06.08493
13 참치 중 수은의 연속 분석을 통해 알아본 수은 분석기 수산물 응용 사례 사이언스212018.05.08507
12 FT-NIR에 의한 유동층건조기에서 분말시료의 수분분석 사이언스212018.05.08207
11 원심분리기 사용자안전 : 바이오안전성 사이언스212018.04.13372
10 Manufacturing of calcium phosphate scaffolds by ps... 사이언스212017.10.10279

비밀번호 확인

게시글 등록시에 입력했던 비밀번호를 입력해 주세요.
기업홍보관 more
세미나
FAQ이용안내
- 실험의 기초와 응용방법
- 시료전처리와 응용방법
- 연구분석장비 응용방법
- 연구장비 세미나,워크숍
- 세계의 연구장비기업들
- 구매공고.제품가격문의
- 전시회.학회.세미나일정
기업홍보관 more
[공지] 사이언스21 매거진,...
[공지] 구정크로마텍 신제품(...
[공지] 사이언스 LAB+ 컨...
[공지] 유나 신제품(1개) ...
[공지] 사이언스21 매거진,...
[공지] 사이언스21 매거진 ...
[공지] 유나 신제품(16개)...
[공지] 길우트레이딩 신제품(...
[공지] 사이언스21매거진,0...
[공지] 구정크로마텍(구정엔지...
[공지] 유나 신제품(9개) ...