Monitoring Drinking Water Quality with the PlasmaQuant® MS
페이스북 트위터 구글 싸이월드 마이피플 
작성자 : 사이언스21 | 조회: 4 | 날짜: 2019-03-11 15:16:10




Introduction
ICP-MS has become the technique of choice for the monitoring of drinking waters. The fast, multi-element capability allows large numbers of samples to be routinely analyzed. While the ability to accurately quantify from ultra-trace to major levels means toxic metals, minerals and other contaminants can be determined within a single measurement. The PlasmaQuant® MS offers fast, multielement and interference-free analysis although with one noticeable difference. Robust plasma conditions suitable for drinking water analysis are achieved with only 7.5L/min of plasma(coolant) gas. The integrated Collision Reaction Cell(iCRC) provides fast and effective removal of spectroscopic interferences using the optimum collision or reaction gas for the job. Allowing the lowest possible detection limits to be achieved for each element. The high pressure/low volume reaction zone also means switching between gas modes takes only seconds.


Instrumentation
A PlasmaQuant® MS with Cetac ASX-520 autosampler and ASXpress rapid sample introduction system was used for the analysis. The Aspect MS software allows automatic optimization of the ion optics, iCRC and plasma gas flows and offers comprehensive quality control protocols, developed to meet the requirements of important regulated methodology including USEPA 200.8.



Samples and Reagents
Calibration standards
Calibration solutions were prepared from high-purity, multi-element solutions in 1% HNO3 covering the concentration range from 0 to 100 μg/L for trace elements and 0 to 100mg/L for major elements.


Sample preparation
NIST Certified Reference Materials 1640; Trace Elements in Natural Water and 1643e; Trace Elements in Water were analyzed directly.


Internal standard
Internal standard contained 1μg/L of Sc, Rh, Ir in 1% HNO3 and was added online via the 3rd channel on the peristaltic pump.


Results and Discussion
Tables 3a and 3b list the average result of triplicate measurements of NIST Certified Reference Materials 1643e and 1640 Trace Elements in Water. Excellent recoveries were recorded and typically fell in the range of ±5%. Analysis times were less than 1.5 minutes per sample using three optimized gas modes: He, no-gas and H2. Stabilization time between gas modes was <5 seconds. Total argon consumption was 9.55L/min which is half the amount typically required on competitive ICP-MS instruments.





Conclusion
This work demonstrates that the PlasmaQuant® MS provides fast and accurate analysis for monitoring the quality of drinking water. The advancement of sample introduction accessories reduces analysis times to less than 1.5 minutes while iCRC technology eliminates problematic interferences quickly and without compromise. A new RF generator design provides robust plasma conditions with only half the argon gas typically required of competitive systems, greatly reducing operating costs and instrument downtime. The capability of the all-digital detector to measure all analyte concentrations from ultra-trace to majors levels in a single ‘pulse-counting’ mode is a major benefit in the analysis of waters. It eliminates the need for performing inaccurate analog measurements when measuring strong signals.


Reference(참고문헌): Analytik-Jena Application note.
Data Services(자료제공): Analytik-Jena Application note.
The Person in Charge(담당자): Choi Jungmi
Maker(제조사): Analytik-Jena
e-mail(이메일):
jung-mi.choi@analytik-jena.com
Country of Origin(원산지): Germany
Model Name(모델명): PQ9000(ICP-OES)


- 2019년 3월호 기사 -



댓글 0
 

댓글쓰기
스팸방지를 위하여 보이는 순서대로 모두 입력해 주세요.
이미지 새로고침

비밀번호 확인

댓글 등록시에 입력했던 비밀번호를 입력해 주세요.

비밀번호 확인

게시글 등록시에 입력했던 비밀번호를 입력해 주세요.

번호제목작성자등록일조회수
55 DSCvesta를 이용한 PET의 유리 전이와 P-Azoxyanisole의 상변화 확인 사이언스212019.03.116
>> Monitoring Drinking Water Quality with the PlasmaQ... 사이언스212019.03.115
53 참치 중 수은의 연속 분석을 통해 알아본 수은 분석기 수산물 응용 사례 사이언스212019.02.1150
52 고분해능 HR-ICP-OES를 이용한 산화아연(ZnO) 분석 사이언스212019.02.1133
51 Application Note # AN232 FT-NIR Quality Control of... 사이언스212019.01.0466
50 데스크탑 X-선 회절분석기의 In-situ 실험 확장 사이언스212019.01.0442
49 CO2 인큐베이터로 효과적인 오염 제어 사이언스212019.01.0446
48 Special Application Note PlasmaQuant® PQ 9000 사이언스212019.01.0424
47 베이비로션과 토양 연속 분석을 통해 알아본 수은 분석기 실생활 응용 사례 사이언스212018.12.0546
46 Impurity Analysis of Zinc Oxides Materials for Ani... 사이언스212018.11.0667
45 DSCvesta를 이용한 PET의 유리 전이와 P-Azoxyanisole의 상변화 확인 사이언스212018.11.0664
44 S6 JAGUAR - full-blown WDXRF analytical performanc... 사이언스212018.10.0875
43 “The Newest XRD Technique” 사이언스212018.10.0867
42 ㈜연진에스텍의 TXA™ Texture analyzer 점착력측정기 사이언스212018.08.29144
41 FTIR microscope 를 이용한 환경시료내 미세플라스틱 분석 사이언스212018.08.06169
40 유리투명전극 기반의 고효율 광전소자 (Highly efficient optoelectron... 사이언스212018.06.0890
39 Micro-XRF를 이용한 국소부위 성분 분석 및 응용사례 사이언스212018.06.0876
38 PlasmaQuant® MS를 이용한 생물학적 매트릭스에서의 요오드 분석 사이언스212018.05.08126
37 PlasmaQuant® MS를 이용한 환경수역에서 수은의 극미량 측정 사이언스212018.04.13174
36 ASBESTOS를 이용한 TG-DTA-PIMS 장비의 동시 측정 응용 사례 사이언스212018.03.09285

비밀번호 확인

게시글 등록시에 입력했던 비밀번호를 입력해 주세요.
- 실험의 기초와 응용방법
- 시료전처리와 응용방법
- 연구분석장비 응용방법
- 연구장비 세미나,워크숍
- 세계의 연구장비기업들
- 구매공고.제품가격문의
- 전시회.학회.세미나일정
기업홍보관
[공지] 사이언스21 매거진 ...
[공지] 사이언스21 매거진 ...
[공지] 지브이에스코리아 신제...
[공지] 지브이에스코리아 신제...
[공지] 사이언스21 2019...
[공지] 한국분석기기 신제품(...
[공지] 글꼴(폰트) 저작권 ...
기업홍보관 more
[공지] 사이언스21 매거진 ...
[공지] 사이언스21 매거진 ...
[공지] 지브이에스코리아 신제...
[공지] 지브이에스코리아 신제...
[공지] 사이언스21 2019...
[공지] 한국분석기기 신제품(...
[공지] 글꼴(폰트) 저작권 ...
[공지] 사이언스21 매거진,...
[공지] 퍼킨엘머코리아 신제품...
[공지] 퍼킨엘머코리아 신제품...
[공지] 사이언스21 매거진 ...
[공지] 싸이스트(주) 신제품...
[공지] 퍼킨엘머코리아 신제품...
기업홍보관 more
세미나
FAQ이용안내