Monitoring Drinking Water Quality with the PlasmaQuant® MS
페이스북 트위터 구글 싸이월드 마이피플 
작성자 : 사이언스21 | 조회: 138 | 날짜: 2018-12-05 12:48:29



Introduction
ICP-MS has become the technique of choice for the monitoring of drinking waters. The fast, multi-element capability allows large numbers of samples to be routinely analyzed. While the ability to accurately quantify from ultra-trace to major levels means toxic metals, minerals and other contaminants can be determined within a single measurement. The PlasmaQuant® MS offers fast, multielement and interference-free analysis although with one noticeable difference. Robust plasma conditions suitable for drinking water analysis are achieved with only 7.5 L/min of plasma(coolant) gas. The integrated Collision Reaction Cell(iCRC) provides fast and effective removal of spectroscopic interferences using the optimum collision or reaction gas for the job. Allowing the lowest possible detection limits to be achieved for each element. The high pressure/low volume reaction zone also means switching
between gas modes takes only seconds.


Instrumentation
A PlasmaQuant® MS with Cetac ASX-520 autosampler and ASXpress rapid sample introduction system was used for the analysis. The Aspect MS software allows automatic optimization of the ion optics, iCRC and plasma gas flows and offers comprehensive quality control protocols, developed to meet the requirements of important regulated methodology including USEPA 200.8.



Samples and Reagents
Calibration standards
C a l ib rat ion s olut ion s we re p rep a re d f r om h i g h - purity, multi-element solutions in 1% HNO3 covering the concentration range from 0 to 100 μg/L for trace elements and 0 to 100 mg/L f or major elements.


Sample preparation
NIST Certified Reference Materials 1640; Trace Elements in Natural Water and 1643e; Trace Elements in Water were analyzed directly.


Internal standard
Internal standard contained 1μg/L of Sc, Rh, Ir in 1% HNO3 and was added online via the 3rd channel on the peristaltic pump.


Results and Discussion
Tables 3a and 3b list the average result of triplicate measurements of NIST Certified Reference Materials 1643e and 1640 Trace Elements in Water. Excellent recoveries were recorded and typically fell in the range of ±5%. Analysis times were less than 1.5 minutes per sample using three optimized gas modes: He, no-gas and H2. Stabilization time between gas modes was <5 seconds. Total argon consumption was 9.55 L/min which is half the amount typically required on competitive ICP-MS instruments.



Conclusion
This work demonstrates that the PlasmaQuant® MS provides fast and accurate analysis for monitoring the quality of drinking water. The advancement of sample introduction accessories reduces analysis times to less than 1.5 minutes while iCRC technology
eliminates problematic interferences quickly and without compromise. A new RF generator design provides robust plasma conditions with only half the argon gas typically required of competitive systems, greatly reducing operating costs and instrument downtime. The capability of the all-digital detector to measure all analyte concentrations from ultra-trace to majors levels in a single ‘pulse-counting’ mode is a major benefit in the analysis of waters. It eliminates the need for performing inaccurate analog measurements when measuring strong signals.


Reference(참고문헌) : Analytik-Jena Application note.

Data Services(자료제공) : Analytik-Jena Application note.
The Person in Charge(담당자) : 최정미 (
jung-mi.choi@analytik-jena.com)
Maker(제조사) : Analytik-Jena
Country of Origin(원산지) : Germany
Model Name(모델명) : ICP-MS (Model : PlasmaQuant MS)

댓글 0
 

댓글쓰기
스팸방지를 위하여 보이는 순서대로 모두 입력해 주세요.
이미지 새로고침
이전글 제약 샘플 분석을 위한 Rigaku장비 소개 사이언스21 2018-07-12 09:42

비밀번호 확인

댓글 등록시에 입력했던 비밀번호를 입력해 주세요.

비밀번호 확인

게시글 등록시에 입력했던 비밀번호를 입력해 주세요.

번호제목작성자등록일조회수
>> Monitoring Drinking Water Quality with the PlasmaQ... 사이언스212018.12.05139
15 제약 샘플 분석을 위한 Rigaku장비 소개 사이언스212018.07.12276
14 생리활성화합물의 작용기전 연구를 위한 융합기술(Integrative Technology f... 사이언스212018.06.08254
13 참치 중 수은의 연속 분석을 통해 알아본 수은 분석기 수산물 응용 사례 사이언스212018.05.08339
12 FT-NIR에 의한 유동층건조기에서 분말시료의 수분분석 사이언스212018.05.08106
11 원심분리기 사용자안전 : 바이오안전성 사이언스212018.04.13129
10 Manufacturing of calcium phosphate scaffolds by ps... 사이언스212017.10.10244
9 데스크탑 X-선 회절분석기를 이용한 In-situ 실험 확장 사이언스212017.09.1416
8 Thrombin Generation in Zebrafish Blood confirmatio... 사이언스212017.09.1414
7 Monitoring Drinking Water Quality with the PlasmaQ... 사이언스212017.09.1420
6 Impurity Analysis of Zinc Oxides Materials for Ani... 사이언스212017.09.1419
5 XRD(X-선 회절분석기), XRF (X-선 형광분석기) 사이언스212017.09.1419
4 FT-NIR에 의한 유동층건조기에서 분말시료의 수분분석 사이언스212017.09.1416
3 광도 측정법에 의한 DNA와 단백질 샘플의 오염도 검출 사이언스212017.09.147
2 Eppendorf 96-Well 세포배양 플레이트- Cell based assay에서 가장... 사이언스212017.09.146
1 HS/GC/MS를 이용한 올리브유 중 BTEXS 분석 사이언스212017.09.145

비밀번호 확인

게시글 등록시에 입력했던 비밀번호를 입력해 주세요.
- 실험의 기초와 응용방법
- 시료전처리와 응용방법
- 연구분석장비 응용방법
- 연구장비 세미나,워크숍
- 세계의 연구장비기업들
- 구매공고.제품가격문의
- 전시회.학회.세미나일정
기업홍보관
[공지] 싸이스트 신제품(2개...
[공지] 싸이스트 신제품(11...
[공지] 싸이스트 신제품(17...
[공지] 싸이스트 신제품(46...
[공지] 싸이스트 신제품(14...
[공지] 싸이스트 신제품(11...
[공지] 구정크로마텍 신제품(...
기업홍보관 more
[공지] 싸이스트 신제품(2개...
[공지] 싸이스트 신제품(11...
[공지] 싸이스트 신제품(17...
[공지] 싸이스트 신제품(46...
[공지] 싸이스트 신제품(14...
[공지] 싸이스트 신제품(11...
[공지] 구정크로마텍 신제품(...
[공지] 싸이스트 신제품(38...
[공지] 사이언스21 매거진,...
[공지] 싸이스트 신제품(17...
[공지] 싸이스트 신제품(19...
[공지] 싸이스트 신제품(14...
[공지] 사이언스21 매거진,...
기업홍보관 more
세미나
FAQ이용안내